0506 009 001 XXX, DEGASS PRESSURE CONTROL ERROR
Unconditional error, fluid path (DIVA/TAVA) is stopped. The blood pump can still be used.
General description of conditions for occurring:
- The flow pump could not create the preset degassing
pressure.
Technical description of conditions for occurring:
- FRI(FI_DegassPr) > (EI(EEI_DegassPr)60)mmHg.
Some possible explanations why the error has occurred:
- DRVA leakage Flow pump/motor broken The preset degassing
pressure |
|