?机型:GE LightSpeed Ultra
故障:打角度时,机架运行缓慢,间断停止
故障信息:File: emcTouchSensor.c Line: 125
Function: PATIENT ANATOMY POSITIONING : Elevation
- Tilt
Scan: 210/3/0 Type: None/Unknown
Exception Level: Pri/Soft Time: 15:18:25:260 Log
Series: 167
The Gantry FORWARD Touch Sensor Error:
Console Pushbutton Remote Tilt is DISABLED.
Use Gantry Controls to Tilt the Gantry.
Current Position 12498 (.01 mm).
Elev Dir: 0 (0 = Up 1 = Down)
(Elev Cmd: 1 TouchSensor: 0x4000).
故障分析:
根据故障信息可以大致分析是Gantry Tilt Touch
Sensors故障,打开机架左右侧盖,仔细检查四个GANTRY OPERATION
BOARD,信号线及电源线连接头紧固,依次拆下各板检测,发现后板Gantry Tilt Touch
Sensors连接不良,处理后正常。 |
|